Wes,
I setup a sheet/Asset Function with the Input range being 240 - 260 Ohms, with the Range Limit -0.0075 to 0.0075 EU ohms. Then the Tolerance is checked on the Output and the Output range is also 240-260 ohms. Tolerance 1 is -0.0075 to 0.0075 EU and If exceeded is Fail. On the Cal Point record is 250 nominal and desired is 250.
While in hindsight I might have messed up the 0.03% to EU, the form seems to accept values/reacts with Pass/Fail as I expected once on a work order.
(I am using MAS Manage 9.0, not Maximo 7.6)
------------------------------
John Q. Todd
Sr. Business Consultant
Total Resource Management
------------------------------
Original Message:
Sent: Tue February 25, 2025 01:41 PM
From: Wesley Ellis
Subject: Calibration - Data Sheet Template for Calibrating a Resistor
I need some help understanding how I would configure a Data Sheet Template for calibrating a resistor. This is the first simple use case for calibration that I am configuring for our Cal Shop. This is a 250Ω resistor and they just check to make sure it is 250Ω ±0.03%. They will adjust the resistance if needed and retire the asset if it cannot be adjusted into spec. This resistor will be used as M&TE for other calibration work orders.
I've tried several different methods of configuring the data sheet, but it seems I have to have an Input and an Output in order to specify a Tolerance. I understand how this is useful for pressure transducers and the like where the input might be PSIG and the output mA, but I don't understand what I should be entering for a single measurement.
------------------------------
Wesley Ellis
------------------------------