Hi Michelle,
We have set up data sheet templates and use repeatable test points on work orders.  It is a very small percentage of our overall data sheet population.  A couple of things regarding repeatable test points that everyone should recognize/remember.
- Tolerance and error calculations against tolerance limits for individual repeated calibration readings are not performed.  These calculations are only performed against the average readings of the repetitions.
- You cannot define tolerance limits for standard deviation, although the value will be calculated by Maximo on the work order.
Specifically for balance calibrations, we normally perform a number of calculations external to Maximo in a spreadsheet.  This spreadsheet is then attached to the work order.  Some values (e.g. standard deviation) from the spreadsheet are transcribed into a data sheet, with defined cal points and tolerance limits, on the work order.
Regards,
Ken
------------------------------
Kenneth Cox
Abbott Laboratories
Abbott Park IL
------------------------------
                                                
					
                                                    
        
                                                
				
                                                
                                                Original Message:
Sent: Fri August 18, 2023 03:06 PM
From: Alvonda Michelle Howard
Subject: Has anyone setup a calibration data sheet to do multiple reading then calculate Standard Deviation, like a balance?
Has anyone setup a calibration data sheet to do multiple reading then calculate Standard Deviation, like a balance?
------------------------------
Alvonda Michelle Howard
------------------------------